• Minimally invasive, scalable cortical microelectrode arrays represent a next step in chronic neural recording and BCI hardware.1
  • The approach combines minimally invasive deployment with scalable fabrication, directly relevant to long-term BCI viability.1 1

Weekly enrichment (2026-07-20)

  • Primary report from Precision Neuroscience: “Minimally invasive implantation of scalable high-density cortical microelectrode arrays for multimodal neural decoding and stimulation,” Nature Biomedical Engineering, 2 October 2025 (doi:10.1038/s41551-025-01501-w; PMID 41039113).234
  • Core device: a 1,024-channel thin-film micro-electrocorticography (µECoG) array comprising 977 recording electrodes (50 µm), 42 stimulation electrodes (380 µm), and 5 reference electrodes (500 µm) at a uniform 400 µm inter-electrode pitch; a 529-channel version (multiple sizes, 300 µm pitch) was fabricated to optimize spacing.5
  • Manufacturing and characterization: process yields were >93% (529-channel) and 91% (1,024-channel); electrode impedance scaled with surface area from ~802 ± 30 kΩ (20 µm) down to ~8.25 ± 0.65 kΩ (380 µm) and was stable before versus after implantation.5
  • Delivery: a “cranial micro-slit” technique makes 500–900 µm skull incisions tangential to the cortical surface, enabling subdural insertion without a burr hole or craniotomy; the full surgical procedure was performed in under 20 minutes.25
  • Preclinical validation: 22 cranial micro-slit insertions across 8 Göttingen minipigs, plus a formal safety/reversibility study in 16 minipigs at 7-day and 42-day timepoints with no observed neurologic impairment; insertion feasibility was also demonstrated in human cadavers.5
  • Scalability and modularity: modules can be tiled — doubly connected 529-channel modules gave 1,058 channels over 0.96 cm² of cortex, and up to four devices (two 529-channel arrays per hemisphere) reached 2,116 channels in a single animal.5
  • Recording and decoding: software allowed real-time visualization of all 1,024 channels, resolving electrocortical activity up to 500 Hz; the system accurately decoded somatosensory, visual, and volitional-walking activity and achieved focal neuromodulation via sub-millimetre cortical stimulation.25
  • Clinical translation: a five-patient intraoperative pilot in anaesthetized and awake neurosurgical patients characterized the spatial scales at which sensorimotor activity and speech are represented at the cortical surface, positioning scalable µECoG as a basis for next-generation BCIs.25

Footnotes

  1. https://news.google.com/rss/articles/CBMiX0FVX3lxTE5RWm84cmZfTXA1ZktRV3gweE91aHliT1dCSC12Z0RKUTZ3QTI4RDlaaG8yazUzbkwzc3F6Zk9jVmdUWExiRi1FRHh4Y0I2WktvdmY5MkxCV3c0VlFYbWlv?oc=5 2 3

  2. https://www.nature.com/articles/s41551-025-01501-w 2 3 4

  3. https://doi.org/10.1038/s41551-025-01501-w

  4. https://europepmc.org/article/MED/41039113

  5. https://www.nature.com/articles/s41551-025-01501-w.pdf 2 3 4 5 6 7